Communication problem Aquadopp via RS232/RS422/wireless network

Up to System integration and telemetry

Communication problem Aquadopp via RS232/RS422/wireless network

Posted by Jeroen Rook at July 10. 2012

Re: Communication problem Aquadopp via RS232/RS422/wireless network

Posted by Jeroen Rook at July 10. 2012

Hi,

 

I am using a Aquadopp profiler, the output is RS232, this is converted to RS422 and Back to RS232.

After that it is connected via MOXA 5400 serial to UTP and over a wireless MESH network.

On the Office site There is another Wireless Mesh receiver with an MOXA 5400 with serial output.

This is connected to the PC with the Aquapro software.

 

Problem is that the Break is not working, if I check in Terminal mode I get the reply : Confirm: with the two reversed P's,

 

What do I have to reply to this Confirm : ? and how can i get it in normal mode ?

 

Can you give me the command set ?

 

Regards,

 

Jeroen Rook

Re: Communication problem Aquadopp via RS232/RS422/wireless network

Posted by Gordon Jones at July 11. 2012
Previously Jeroen Rook wrote:

Hi,

 

I am using a Aquadopp profiler, the output is RS232, this is converted to RS422 and Back to RS232.

After that it is connected via MOXA 5400 serial to UTP and over a wireless MESH network.

On the Office site There is another Wireless Mesh receiver with an MOXA 5400 with serial output.

This is connected to the PC with the Aquapro software.

 

Problem is that the Break is not working, if I check in Terminal mode I get the reply : Confirm: with the two reversed P's,

 

What do I have to reply to this Confirm : ? and how can i get it in normal mode ?

 

Can you give me the command set ?

 

Regards,

 

Jeroen Rook

Hi Jeroen,

 

It sounds like your ADCP might already be taking measurements.  You should try clicking "stop recorder deployment", then send a break.  This should return the ADCP model etc as per usual.  All the commands can be found in the user interface manual at:

 

http://www.nortek.no/en/support/application-development

 

Regards

 

Ramsay

Re: Communication problem Aquadopp via RS232/RS422/wireless network

Posted by Cees Meijer at July 13. 2012

Hi Jeroen,

 

Just type 'MC' after you get 'Confirm' and your Aquadopp will enter command mode.

 

Cees

 